Re: GFP and YFP on the LSRII/Aria

From: Alexandre Bote Tronchoni <cafeoleman@gmail.com>
Date: Fri Jul 04 2008 - 04:44:28 EDT
Hi Mark,

We are using new filters in our FacsAria to perform EGFP-YFP separation.

For EGFP we use a 495LP and a 510/20 BP
For YFP we use a 527LP and a 550/30 BP

We use a compensation about 20-30% for each one.
The results are quite good.
We acquire the filters at Omega Filters.

Alexandre Bote Tronchoni
UCTS - Unitat Científico-Tècnica de Suport
Plataforma de Citòmica
Telf: +34 934894179 Fax: +34 934894185
Institut de Recerca Hospital Universitari de la Vall d'Hebron
Passeig de la Vall d'Hebron 119-129
08035 Barcelona
http://www.ir.vhebron.net/easyweb_irvh/default.aspx?language=ca


Si té solució, de què et queixes?
Si no té solució, de què et queixes?

On Thu, Jul 3, 2008 at 14:35, Mark Shlomchik <mark.shlomchik@yale.edu>
wrote:

> Hi All,
>
> Has anyone been separating GFP and YFP on the LSRII/Aria?
>
> I am interested in suggestions for filter sets that will minimize
> compensation and maximize sensitivity (aren't we all?) In any case, if you
> can share your experience that would be helpful before we order filters.
>
> Thanks,
>
> Mark
> --
> Mark Shlomchik, MD, PhD
> Professor of Laboratory Medicine
> and Immunobiology
> Yale University School of Medicine
>