From: Bushnell, Timothy <tim_bushnell@URMC.Rochester.edu>
Date: Thu Jul 03 2008 - 16:49:42 EDT
Mark:

We have been doing this extensively on both an LSR-II and FACSAria in
multichromatic work (8+ colors) For example, Neering et al., Blood
110:2578-2585 (2007)
(http://bloodjournal.hematologylibrary.org/cgi/content/full/110/7/2578)

We use:
	LP	BP
YFP	525LP	530/30
GFP	502LP	510/20

Regards
Tim

Tim Bushnell, Ph.D.
Research Assistant Professor, Pediatrics and Oncology
Co-Director, URMC Flow Cytometry Facility
Office: 585-273-5535
Lab: 585-273-1361
Cell: 585-690-5157
Fax: 585-276-0233
http://www.urmc.rochester.edu/flow-core/
http://www.urmc.rochester.edu/wnyfug/



On 7/3/08 8:35 AM, "Mark Shlomchik" <mark.shlomchik@yale.edu> wrote:

> Hi All,
> 
> Has anyone been separating GFP and YFP on the LSRII/Aria?
> 
> I am interested in suggestions for filter sets that will minimize
> compensation and maximize sensitivity (aren't we all?) In any case,
> if you can share your experience that would be helpful before we
> order filters.
> 
> Thanks,
> 
> Mark
Received on Fri Jul 4 19:38:00 2008