I agree that there's virtually no way to reliably measure these  
volumes you are considering. Another method perhaps to accomplish  
essentially the same thing would be to use counting beads of some  
sort (Bangs labs and Invitrogen for example carry these; BD has  
Trucount tubes for this purpose also.)
The advantage is that using beads of a known concentration, once you  
add beads to a sample in known proportions (ie.  3:1  or 1:1) and  
count a reliable number of beads to make the count statistically more  
accurate, your	estimate the concentration or absolute count of your  
cells relative ot the beads regardless of fluidic volumes, additional  
stuff added to the tubes , etc. This will allow you the single	
platform count determination that you're looking for with a much more  
reliable method.

Wayne A. C. Harris
Bone Marrow Transplant Laboratory
Immunology and Flow Cytometry Assays Laboratory
Winship Cancer Institute, Emory University
1701 Uppergate Drive
WCI, building C, Rm#4032
Atlanta GA  30322
404-727-3086



>
>
> Sergei Ochkur wrote:
>>
>> Can we count absolute cell number/concentration by simply dividing  
>> total cell count (or number of events collected) by the volume  
>> passed through the flow cytometer?
>>
>> The volume here is:
>> (volume of counted cell suspension) = (volume in the tube before  
>> flow) - (volume in the tube after flow) - (volume of the dead  
>> space the flow Cytometer lines between the intake and the  
>> measuring chamber)
>>
>> If yes, how can we find/measure that dead space volume?
>>
> With difficulty. And there are some other problems; you can't  
> estimate the volume in the tube all that accurately, even if the  
> tube is graduated, and the overall strategy assumes that there is  
> never any sheath blown back into the tube during sample handling,  
> and also that there is no instrument dead time, which may apply to  
> some all-digital systems but does not to older flow cytometers.
>
> A method for getting reasionably accurate counts from cytometers  
> with stable flow rates
> was described in:
> Bergeron M, Lustyik G, Phaneuf S, Ding T, Nicholson JK, Janossy G,  
> Shapiro H, Barnett D,
> Mandy F. Stability of currently used cytometers facilitates the  
> identification of pipetting errors
> and their volumetric operation:
> "time" can tell all. Cytometry B Clin Cytom. 2003; 52:37-9 (free  
> download online).
> The real question is "how good is good enough"? Hematology  
> instruments, which are flow
> cytometers, have been
> designed to yield accurate and precise counts; manufacturers have,  
> for the most part, not
> provided the same capability in
> flow cytometers intended for other purposes, including clinical  
> immunophenotyping.
>
> -Howard
>
>
>
>