RE: Aria and 130 nozzle

From: Greg Veltri <Greg.Veltri@cancer.org.uk>
Date: Sat May 10 2008 - 04:30:49 EDT
Hi Barb,

I have a 130um nozzle and using a frequency between 17-18 at 12 psi works well for us.  We can see the break-off and have no issues sorting.  You may need to place the attenuation on.  The key is to make sure you get the air out of system (let it sit 30 -60 mins before accudrop) and start with a full PBS container.  

Good luck,

Greg