HI Anne - it's not necessarily the answer in your case, but it would fit the reported results. The far-red sensitive PMT's fitted in BD and MoFlo machines can give large numbers of spikes (due to stray scattered laser or thermal photons) when the voltage is turned up high. If the spikes coincide with a negative cell going through the machine then it appears as a dimly positive event, negative cells that go through out of sync with the single-photon or thermal spikes stay negative. It may be the case that you have the PMT's turned up for your cell sorts, then down much lower when sorting beads. Sorting these (negative cells) then reanalyzing of course gives the same result (negative cells, the same fraction of which go through at the time of a spike). You can test this quite easily, turn the PMT to the same voltage used for sorting and trigger on that parameter (you'll probably get a normal distribution midway up the graph when triggering just above zero), or just check the oscilloscope trace for that parameter when running blank beads (also check the distribution of blank beads with the same PMT settings). Hope this makes sense and helps - if not drop me a line and I'll send you some example plots, Cheers Ray Ray Hicks Cytek Development Inc http://www.cytekdev.com tel: +44 (0)208 1337 968 (UK/Europe) tel: +1 510 931 7294 (USA) skype: ray.hicks.cytek From: Ann Atzberger [mailto:Atzberger@hammer.imm.ox.ac.uk] Sent: Monday, April 07, 2008 7:52 AM To: cyto-inbox Subject: CD34+CD38- Hi everybbody, recently we have been sorting CD34Fitc positive CD38PerCPCY5.5 negative BM cells on a MoFlo. This population although correct in number of recovered cells shows no enrichment whatsoever on reanalysis. This population is part of a 4-way sort. The other 3 populations when reanalysed have purities above 95%. (Total colours = 6) We switched to using PE-CY5 instead of PerCP5.5. and saw the same problem. Using exactly the same voltage and compensation settings we did a 4-way sort with beads and got 100% purities on all populations...which rules out malfunction of the instrument. We reverted back to labelling the CD38 with Fitc and CD34 on PE and found that the purities were back to normal i.e. above 90%. Has anyone come across a similar problem with tandem conjugates? I am reluctant to say that aCD38 cannot be used with a tandem conjugate label and would greatly appreciate any plausible explanation for this phenomenon. We are very much aware of the instability of tandem conjugates and the circumstances that exacerbate this, the tandems mentioned are purchased from Beckman Coulter - as are most of the tandems we work with and on the whole our experience with them is good. regards AnnReceived on Tue Apr 8 13:18:00 2008
This archive was generated by hypermail 2.1.8 : Wed Jan 31 2007 - 03:12:00 EST