RE: Doublets?

From: Nebe-Von-Caron, G <g.nebe-von-caron@spdspark.com>
Date: Mon Feb 11 2008 - 12:02:08 EST
Gates or no Gates...


For a long time I have tried to convince the flow community to look at
raw log side scatter versus the various fluorescence channels,
particularly when looking at bacteria or nano and microparticles. 

Cells with different scatter properties can be an lysis artefacts
coincidences or functional cell interactions, blasts.... After all we
teach everyone that the cells hug each other for signalling. Therefore
it is essential to check the untampered data, ideally looking at log
side scatter versus log forward scatter and also to observe the log side
scatter versus log fluorescence data. 

The degree of aggregation can vary depending on the anticoagulant used,
sample age, temperature,  the lysis method, the number of centrifugation
steps... and the patient!!! Certain disorders show very selective
cell-cell adhesion.

Find enclosed an example were I just pulled up the lin and log scatter
pictures (apologies for the orientation to the ones who think the other
way round and that the area data are out of scale) and the log side
scatter/CD3 picture. Gating on the CD3+ monocytes instantly updates the
preview plane from which I then selected the CD4 vs. cd8 picture to show
the biased aggregation of cd3+ cells towards CD8+ which shows it neither
to be coincidence or random aggregation. The monocyte and
lymphocyte-monocyte aggregates can even be separated by area forward
scatter (yes it would be nicer if on scale).

The power of flow cytometry lies in the multiparametric correlation of
data. Whilst the VenturiOne software offers a very fast and convenient
way of seeing all the correlations at once there is no excuse for not
generating at least the raw side scatter versus fluorescence plots on
any conventional software package with either Gates - or Jobs
technology.




Gerhard Nebe-von-Caron 
Research Scientist and Biomedical Engineer 
SPD-Spark
Swiss Precision Diagnostics 
Priory Business Park 
Bedford, MK44 3UP, UK 
Tel +44(0)1234-835474 
Fax +44(0)1234-835002 
mailto:g.nebe-von-caron@spdspark.com 

-----Original Message-----
From: Sylvie Amu [mailto:sylvie.amu@rheuma.gu.se] 
Sent: 05 February 2008 22:25
To: Cytometry Mailing List
Subject: Re:Doublets?

Halo

We do see the same phenomenon running our samples in CantoII.  We work
with human blood
and are staining for B cells (CD19) and T cells (CD3) and see a
population at the same
location you are describing being double positive CD19+CD3+, doublets.
This population
will disappear if you gate on single cells firs ( FSC-A vs FSC-H) and
then continue your
analyses. This has been our approach.  Hope it helps.

Regards 

Sylvie Amu, PhD student
Dept of Rheumatology and Inflammation Research 
Gothenburg University
Sweden




On Feb 1, 2008, at 7:14 AM, Rudensky, Bernard wrote:

Many times we find a second small population on the fsc vs ssc plot
which has the ssc
properties of lymphocytes but higher fsc values approaching that of
monocytes. By
fluorescent markers this population appears to be lymphocytes with the
same phenotype as
the major lymphocyte population. We do not see this with every assay but
the phenomenon
seems to be experiment dependent as every tube in the same experiment
gives the same
picture.
The Canto technician thinks that the second population represents
doublets.
Does anyone have experience with this phenomenon? Is there a way to
avoid it? Can it
affect diagnostic results?
Your input would be most appreciated.
Professor Bernard Rudensky
Clinical Laboratories
Shaare Zedek Medical Center
Jerusalem, Israel 91031
IMPORTANT: The contents of this email and any attachments are
confidential. They are
intended for the named recipient(s) only.
If you have received this email in error, please notify the system
manager or the sender
immediately and do not disclose the contents to anyone or make copies
thereof.
*** eSafe scanned this email for viruses, vandals, and malicious
content. *** Shaare
Zedek Medical Center http://www.szmc.org.il





This attachment - 'doublets.jpg' -  653.79 KBytes - can be viewed at
http://www.cyto.purdue.edu/MD-parts/9259562d6cf9f16209bc0499ba3055ac5286bb12.jpg 
Received on Tue Feb 12 14:58:00 2008