FW: 6 side streams in ARIA?!?

From: Pizzo,Eugene <Pizzo@NSO1.UCHC.EDU>
Date: Fri Dec 07 2007 - 10:08:32 EST
Gene Pizzo, M.S
Manager, Flow Cytometry Facility
UCONN Health
Farmington, Ct. 06030
860 679-7567
http://flowcytometry.uchc.edu <http://flowcytometry.uchc.edu/> 

________________________________

From: Pizzo,Eugene
Sent: Thu 12/6/2007 18:37
To: cyto-inbox
Subject: RE: 6 side streams in ARIA?!?


Hi Vinko,
That's a really cool picture and in some sense it shows the future of flow cytometry as
the engineers learn to add more streams AND more colors to sorting but what you have is
not six streams but four streams with center stream fanning. It is the result of a drop
drive frequency that is "inadequate" to clarify drop separation. 
On the ARIA large nozzle sizes are somewhat unstable and require a fine tuning of nozzle
position, drop drive frequency, amplitude, and pressure. Use the built-in "low" setting
only to ballpark. Then begin by adjusting pressure and drop drive frequency to be certain
you're getting good separation of drops and merging satellites.
Because the separation of your streams looks otherwise so good I'd guess it's just a
minor frequency change to tighten your center stream. You're plate voltage - at 5000
volts - is too high for a large nozzle however, that could be "stretching" your streams.
I only use such a high voltage for higher pressures/smaller nozzles. I run the 100um
nozzle closer to 2,500 volts. Don't bother with the 2nd, 3rd, and 4th drop adjustment -
strangely I've never seen this have any effect since the old Vantage/SE days - since
we've gone digital if you need to use the drops adjustment it's a good bet you're in the
wrong frequency. Keep it at 20,10,5.
Gene Pizzo, M.S
Manager, Flow Cytometry Facility
UCONN Health
Farmington, Ct. 06030
860 679-7567
http://flowcytometry.uchc.edu <http://flowcytometry.uchc.edu/> 

________________________________

From: Vinko Tosevski [mailto:vinko.tosevski@medri.hr]
Sent: Wed 12/5/2007 17:11
To: cyto-inbox
Subject: 6 side streams in ARIA?!?



Hi,

My department recently bought an Aria and I was among the first ones to go
through operator's training for it. So, although I have previous flow
experience, I am largely a beginner in the sort world.

Today I tried to sort some cells using 100u nozzle and LOW sort mode (for
the first time; most of the time we use 70u nozzle and HIGH sort mode). The
stream looked OK but, when I applied the voltage and used TEST SORT button,
I got 6 side streams instead of 4!??! (I will try to attach a picture but
I've never seen any attachment going through) Is there a simple explanation
to this phenomenon and how can I resolve this? When I move the sliders used
to adjust the position of side streams, the outer two side streams on each
side respond. But, the center stream splits/gets 2 side streams of its own
after applying a voltage, and they are inresposive to the movement of any
sliders. These two are not too much apart but are clearly visible as
separate streams in camera window...

Any comment is more than welcome.

Sincerely,
Vinko

----
Vinko Tosevski, MMolBiol

Research fellow/Teaching assistant,
Department of physiology and immunology,
Rijeka Medical School,
Ul. b. Branchetta 20,
HR-51000 Rijeka, Croatia

Tel: +385 51 651 192
Fax: +385 51 675 699
Received on Fri Dec 7 14:38:00 2007