Hi Albert, if you don't like the Aria fluidics cart compressor (it does the job but is rather noisey), you can easily hook up any external clean pressurized air supply to the cart via a reinforced hose and a CPC coupling. The build-in compressor is then switched off via a switch on the cart's side. You can get the right kind of hoses from any hardware store. The coupling I got from the local CPC dealer, but BD can help you find the right one, or even supply it if you ask nicely. I do recommend installing a high-quality oil filter (two-stage) if you take air from an oil compressor. We were surprised just how much oil accumulates from very fine aerosol over failry short periods of time if you run standard filtered air from such a compressor through liquid (which may not be the case in the Aria, but I'm bringing this up because the film builds up visibly on top of the liquid - which is easy to detect). Guy Next generation therapeutic <http://www.ablynx.com/> antibodies Guy Hermans, PhD Senior Scientist Ablynx NV Technologiepark 4 B-9052 Zwijnaarde Belgium guy.hermans@ablynx.com tel: fax: mobile: +32 (0)9 261 06 57 +32 (0)9 261 06 27 +32 (0)486 788 551 <https://www.plaxo.com/add_me?u=30065269879&v0=994426&k0=2009290972> Add me to your address book... <http://www.plaxo.com/signature> Want a signature like this? -----Original Message----- From: Albert Tai [mailto:acktai@exelixis.com] Sent: Tuesday, April 24, 2007 7:39 PM To: cyto-inbox Subject: Air compressor for FACS Aria Hi Flow-ers, I am wondering if anyone has experience of adding a JUN-AIR compressor to their FACS Aria ? I talked to BD and they recommended model 6-25 but the tech support could not tell me what other parts I would need to connect the compressor to the Aria nor would he advise a protocol for making that connection. Your comments on part numbers and/or protocol would be much appreciated. Thanks in advance. Albert Exelixis, Inc. This email (including any attachments) may contain material that is confidential and privileged and is for the sole use of the intended recipient. Any review, reliance or distribution by others or forwarding without express permission is strictly prohibited. If you are not the intended recipient, please contact the sender and delete all copies. Exelixis, Inc. reserves the right, to the extent and under circumstances permitted by applicable law, to retain, monitor and intercept e-mail messages to and from its systems. ----------------------------------------------------------------------- THIS E-MAIL MESSAGE IS INTENDED ONLY FOR THE USE OF THE INDIVIDUAL OR ENTITY TO WHICH IT IS ADDRESSED AND MAY CONTAIN INFORMATION THAT IS PRIVILEGED, CONFIDENTIAL AND EXEMPT FROM DISCLOSURE. If the reader of this E-mail message is not the intended recipient, you are hereby notified that any dissemination, distribution or copying of this communication is strictly prohibited. If you have received this communication in error, please notify us immediately at ablynx@ablynx.com. Thank you for your co-operation. "NANOBODY" and "NANOCLONE" are registered trademarks of Ablynx N.V. -----------------------------------------------------------------------
This archive was generated by hypermail 2.1.8 : Wed Jan 31 2007 - 03:12:00 EST