From: <akos.szilvasi@novartis.com>
Date: Wed Apr 18 2007 - 15:57:57 EDT
Hi, Ultan,

You must have steady temperature for the MoFlo. That sorter is sensitive 
to the room temperature changes. The drop delay will move, the stream 
elongates as the room warms up....and the room will warm up when you 
operate them because of the heat output of the different components. Make 
sure  - if you need to have water cooled lasers - that the heat exchanger 
is outside of the flow room. However, a 40 sm space has a big temperature 
buffering capacity that may not cause problems.

Best regards,

Akos

__________________________ 
Akos Szilvasi
NIBRI Core Laboratory Services
manager
USCA, 601-5301
Novartis Institutes for BioMedical
Research, Inc.
100 Technology Square
Cambridge, MA 02139
USA
Phone: +1 617 8717177
Email : akos.szilvasi@novartis.com






Ultan Cronin <ultan.cronin@ul.ie> 
04/18/2007 03:32 PM

To
Cytometry Mailing List <cytometry@flowcyt.cyto.purdue.edu>
cc

Subject
Air Conditioning for Sort Lab






Dear Flowers,

We will soon be installing an air conditioning system in our lab, which 
contains one BD LSR I and a MoFlo.  The lab has one door, three large 
windows and would be about 40 meters sq.  Has anyone any advice to offer 
us before we commit a design to paper?

Yours gratefully,

Ultan
-- 
Ultan Cronin BSc MSc
Dept of Life Sciences
Schrodinger Building
University of Limerick
Castletroy
Co. Limerick
Ireland
Tel: 00353-(0)61202890
Mob: 00353-(0)872032963
Fax: 00353-(0)61331490

_________________________

CONFIDENTIALITY NOTICE

The information contained in this e-mail message is intended only for the 
exclusive use of the individual or entity named above and may contain 
information that is privileged, confidential or exempt from disclosure 
under applicable law. If the reader of this message is not the intended 
recipient, or the employee or agent responsible for delivery of the 
message to the intended recipient, you are hereby notified that any 
dissemination, distribution or copying of this communication is strictly 
prohibited. If you have received this communication in error, please 
notify the sender immediately by e-mail and delete the material from any 
computer.  Thank you.


picture
Received on Thu Apr 19 17:18:02 2007

This archive was generated by hypermail 2.1.8 : Wed Jan 31 2007 - 03:12:00 EST