RE: A big thank you on biosafety

From: Badore, Leslie C. <Leslie.Badore@vtmednet.org>
Date: Tue Sep 19 2006 - 10:22:02 EDT
Dear Petra and Joern -
 
Here in our clinical flow cytometry lab, we use 10% bleach, followed by
distilled water, for cleaning. We generally use one tube with 10%
bleach, followed by 3 tubes of distilled H2O  All are run for 200
seconds each. We also use 10% bleach and DH2O for vaccuum cleansing.  It
is not necessary to use undiluted bleach.  We only use undiluted bleach
for a large clog or other troubleshooting - it will affect your lines.
The most important point is that the 10% bleach needs to be made fresh
daily.
 
We run T cell subsets on a large population of HIV positive patients.
We use a lysing system that inactivates or kills the HIV virus.  We also
always observe standard precautions.  You can also get cleaning agents
that are specifically made for cleaning in potentially biohazardous
areas (such as a laboratory).  
 
These were the instrument manufacturers' instructions.	They have worked
very well for us for about 15 years now.  Indeed, better safe than
sorry, although bleach stronger than 10% will not be any more effective.
 
Hope this helps. 
 
Best regards - Leslie	 
 
Leslie C. Badore, CLS(NCA)
Immunology Charge Technologist
Pathology & Laboratory Medicine, EP1-100
Fletcher Allen Health Care
111 Colchester Avenue
Burlington, VT 05401
email: Leslie.Ryan@vtmednet.org
phone: (802)847-5373
fax: (802)847-4103
 


Confidentiality Notice:
This message, and any attachments, may contain information that is confidential, privileged, and/or protected from disclosure under state and federal laws that deal with the privacy and security of medical information. If you received this message in error or through inappropriate means, please reply to this message to notify the Sender that the message was received by you in error, and then permanently delete this message from all storage media, without forwarding or retaining a copy.
Received on Tue Sep 19 13:58:00 2006

This archive was generated by hypermail 2.1.8 : Wed Sep 20 2006 - 03:12:04 EDT