Let me add a note of caution regarding the use of pyronin Y as a marker of proliferating cells or RNA. Both Howard and Edward are right that a bivariate analysis of Hoechst vs pyronin Y fluorescence allows one to discriminate between Go vs G1 cells. Edward is also correct stating that pyronin Y is toxic to cells. I would argue, however, that when used supravitally pyronin Y preferentially stains mitochondria and not RNA (see e.g. "Cytostatic and cytotoxic properties of pyronin Y: Relation to mitochondrial localization of the dye and its interaction with RNA", Cancer Res., 46: 5760-5766, 1986). In fact, the cytotoxic and cytostatic properties of pyronin Y remind very much rhodamine 123 and relate to its interaction with mitochondria. The cytotoxicity is enhanced when cells stained with pyronin Y are illuminated with white light - e.g. under light microscopy one can easily see that mitochondria containing pyronin Y swell and rupture within seconds after turning on light. Noncycling (Go) cells, have much fewer mitochondria and their stainability with Rhodamine 123, similar as with pyronin Y, is many-fold lower that that of G1 -S- G2M lymphocytes ("Increased mitochondrial uptake of Rhodamine 123 during lymphocyte stimulation" Proc. Natl Acad. Sci. USA 78: 2383-2387, 1981) The photosensitizing effects of pyronin Y through mitochondria was described in another publication ("Photosensitizing effect of tricyclic heteroaromatic cationic dyes pyronin Y and toluidine blue O (Tolonium chloride" Cancer Res., 48: 1295-1299, 1988)"). Perhaps to lower the cytotoxicity of pyronin Y one has to carefully screen cells from the ambient light. Rhodamine 123 may be less toxic than pyronin Y and both discriminate Go from G1 cells. Of course, pyronin Y and Hoechst can be used to stain differentially RNA and DNA in fixed cells. Pyronin Y can also be used to discriminate between ds and ssRNA ("Application of pyronin Y in cytochemistry of nucleic acids". Cytometry, 8:138-145,1987). Happy New Year to all FLOWERS !!!!! Zbigniew Darzynkiewicz, M.D., Ph.D. Brander Cancer Research Institute at NYMC 19 Bradhurst Avenue, Suite 2400 Hawthorne, N.Y. 10532 darzynk@nymc.edu http://www.darzynkiewicz.com/zbigniew/ -----Original Message----- From: Edward F. Srour [mailto:esrour@iupui.edu] Sent: Tuesday, December 27, 2005 10:50 AM To: cyto-inbox Subject: Proliferating lymphocytes in peripheral blood and CD25 expression This exchange took place between Petros Christopoulos and Howard Shapiro Petros Christopoulos wrote: >I would like to ask if anybody knows about the relationship of >expression of classical >proliferation markers (like Ki-67 or the transferrin receptor) and >CD25 (which is also an activation marker) in peripheral blood lymphocytes. >Do they give similar results, do they correlate or are they totally >independent of each >other ? Time to start another winning streak for Mailing List submissions... See Figure 10-7, p.460, in the 4th Edition of Practical Flow Cytometry (available without charge on the Invitrogen/Molecular Probes web site, http://probes.invitrogen.com/) for plots relating transferrin receptor (CD71) expression and RNA content to DNA content. I have speculated in the book and on this Mailing List that Ki-67 and CD71 should correlate almost absolutely in proliferating lymphocytes and probably in most or all other cell types, but, when I just looked on PubMed, I could not find any publication suggesting that both had been measured simultaneously in the same cells at the same time. This really needs to be done; there have been over 100 papers published referring to Ki-67 in the past year, representing a lot of work by many people. Detecting Ki-67 requires permeabilization and is only possible in fixed cells, whereas detecting CD71 requires only surface staining, meaning that live cells can easily be analyzed and sorted based on CD71 status, but not based on Ki-67 status. If the two markers are as well correlated as I suspect, a switch from Ki-67 to CD71 staining will save both time and money, and, more importantly, allow researchers to do further functional studies on live cells precisely characterized as to cell cycle position. Such studies can, in at least some cases (see the work of Edward Srour et al), be done using Hoechst 33342/Pyronin Y staining for DNA and RNA, but pyronin Y can be toxic. _________________________________________________________ So, my input on this topic is the following. Howard's assumption is right on the money. expression of CD71 and Ki67 correlate well with one another. we have isolated murine cells lacking the expression of CD71 versus cells that are CD71+. upon PCR analysis of these cells, we found that CD71- cells do not express Ki67 while CD71+ cells do. the separation between CD71+ and CD71- cells is a little tricky depending on the cell type since many cell groups present a continuum of CD71 staining such that separation between positive and negative is rather subjective. As others argued on this forum that an isotype control is in general useless, this may be a case where the use of an isotype may be helpful. we have not published this yet. Also another point of agreement with Howard. Pyronin Y is toxic especially to murine cells while it is well tolerated by human cells. unfortunately, some people are publishing the use of Pyronin Y with murine hematopoietic cells without recognizing that although the data presented clearly indicate something wrong was going on in these cells. Happy Holidays edward f. srour Edward F. Srour, Ph.D. Professor of Medicine, Pediatrics, Micro/Immunol. Indiana University School of Medicine Cancer Research Institute 1044 West Walnut Street. R4-202 Indianapolis, IN 46202-5121 Phone: 317-274-0343 317-274-3589 Fax: 317-274-0396 email: esrour@iupui.edu -- End --Received on Wed Dec 28 14:38:00 2005
This archive was generated by hypermail 2.1.8 : Sat Jan 14 2006 - 22:04:01 EST