RE: Bacterial enumeration: beads versus CFU

From: Nebe-Von-Caron, G <g.nebe-von-caron@unipath.com>
Date: Thu Dec 01 2005 - 13:29:53 EST
The most common discrepancy between CFU's and cytometric counts
originates from identifying non-bacterial events as bacteria e.g.
counting noise or bacteria originating from buffers etc but not from the
sample. The use of DNA stains can help to identify what is bacteria and
what is noise it can be misleading if the bacteria possess transport
systems to chuck out the DNA stains. Other problems are aggregation and
coincidence of events that need to be considered. Once those potential
interferences are sorted out things should not be too bad. However there
are systematic problems originating from the term "viability". 
 
A staining based on DNA stains with differential membrane permeability
can only indicate the permeability of the membrane to the respective
probes, so you measure membrane integrity. Whilst in general cells
taking up PI would be classified as dead permeabilisation can be
transient as under conditions of electroporation, chemoporation etc.
 
Nonpermeabilised cells have at least the potential to form gradients but
do not necessarily do so. Thus a discrepancy between reproductive viable
cells and intact cells has to be expected. In fact if you irradiate
bacteria of treat them with strong oxidative stresses you can generate
bacteria that show membrane integrity, even metabolic activity but still
won't reproduce. 
 
If you look at perfectly happy marine samples it is generally accepted
that even if the cells show metabolic activity, the number of cells
growing in culture is still only a small fraction of the cells
detectable as they are very fastidious. Similarly if you look at
stressed or injured cells most culture based recovery methods fail. The
discrepancy between different recovery methods can give you up to 5 log
difference. In the light of those problems even the differentiation
system described in
http://www1.elsevier.com/homepage/sah/mimet/speciss/1378.pdf does not
help to consolidate the data.
 
Overall you should expect the numbers of intact cells to exceed the
number of reproductive viable cells if measurements are done correctly,
as the reproductive growing cells are a subpopulation of the metabolic
active cells which in turn are a subpopulation of the intact cells.
 
regards
 
Gerhard

	-----Original Message-----
	From: Therese M. McGinn [mailto:tmcginn@benchmarkbiolabs.com] 
	Sent: 28 November 2005 16:16
	To: Cytometry Mailing List
	Subject: Bacterial enumeration: beads versus CFU
	
	

	Hello All,

	 

	I have been using the Becton Dickinson Cell Viability Kit with
Counting Beads to determine numbers of live bacteria in cultures.  This
method involves staining with propidium iodide and thiazole orange to
discriminate live from dead cells.  Numbers of live cells are determined
by a simple calculation relating the number of events in the live gate
to the number of beads.  

	 

	We have been comparing the numbers of live cells by this method
to CFU/ml as determined by traditional enumeration assay.  The
flow-based method shows much higher numbers of live cells than the
enumeration assay.  I would appreciate any feedback or suggestions on
how to reconcile this discrepancy.  

	 

	Thank you for your consideration,

	 

	Terry McGinn

	 

	 

	Therese M. McGinn Ph.D.

	Senior Research Scientist

	Benchmark Biolabs, Inc.

	Phone: 402-475-8104

	Fax: 402-475-8511

	www.benchmarkbiolabs.com

	 

	
	
	
**********************************************************************
	This email and any files transmitted with it are confidential
and
	intended solely for the use of the individual or entity to whom
they
	are addressed. If you have received this email in error please
notify
	the system manager.
	