Hi Larry, We use these two all the time without any problem, and the only thing I can see that's different about your setup is the 685LP. I'm thinking it may be too long- it looks like you're catching it right at the peak of Cy5.5 emission, so you're directing a lot of Cy5.5PE light into the PE detector. Our setup is custom, and probably the reverse of yours, but we have a 630SP that divides the light one way for PE and PETR detection, and the other way for Cy5PE, Cy5.5PE, and Cy7PE. Have you tried anything shorter than 685? I just looked back, and we also used this filter when we had the standard Z configuration. We've always run 200 mW 488 nm, a 580/30 for PE, and a 720/45 for Cy5.5 PE. Hope that helps! Carol > > I am having a problem on my MoFlo that I can not figure out. When I run > PE-Cy5.5 I am getting a huge amount of signal in the PE channel (or > maybe > too little in the PE-CY5.5 channel). So much that I can not compensate > properly - needs about 60-70% and then still doesn't look right. When > I run > the same sample on the FACSCalibur it looks great and only needs 12% > compensation. Thus the problem is not the reagent. I have considered > the > filters as a problem but 2 different PE filters gave the same result. I > have used a 580/30 and a 575/25. These filters in front of the PE > detector > in all other fluorochrome combos that I have tried give no problem. > Other > parameters are 200mW 488nm, 685LP dichroic and a 700/50 filter in > front of > the PE-Cy5.5 detector. I have not tried other filter > combos in front of > the PE-Cy5.5 detector. Anybody have any ideas? Thanks > > Larry > > > Larry W. Arnold, Ph.D. > Research Professor and Director, Flow Cytometry Facility > Department of Microbiology and Immunology > Lineberger Comprehensive Cancer Center > CB# 7290 > University of North Carolina at Chapel Hill > Chapel Hill, NC 27599 > Phone: 919-966-1530 > FAX: 919-962-8103 > > > > <image.tiff> > > > ******************************************************************* > IMPO RTANT MESSAGE FOR RECIPIENTS IN THE U.S.A.: > This message may constitute an advertisement of a BD group's products > or services > or a solicitation of interest in them. If this is such a message and > you would like to opt out of receiving future advertisements or > solicitations from this BD group, please forward this e-mail to > optoutbygroup@bd.com. > ******************** *********************************************** > This message (which includes any attachments) is intended only for the > designated recipient(s). It may contain confidential or proprietary > information and may be subject to the attorney-client > privilege or other confidentiality protections. If you are not a > designated recipient, you may not review, use, copy or distribute this > message. If you receive this in error, please notify the sender by > reply e-mail and delete this message. Thank you. > ************* ****************************************************** > Corporate Headquarters Mailing Address: BD (Becton, Dickinson and > Company) 1 Becton Drive Franklin Lakes, NJ 07417 U.S.A. > ******************************************************************* > > ------------------------------------------------------------------------ --- Carol Oxford Manager, UC Davis Optical Biology Lab Tupper Hall, rm 3425 University of California Davis, California 95616 (530) 752-7205 (530) 752-4548 fax cloxford@ucdavis.edu ------------------------------------------------------------------------ ---Received on Thu Jun 2 15:18:01 2005
This archive was generated by hypermail 2.1.8 : Sat Jan 14 2006 - 22:03:47 EST