Dear Achal, To solve this problem, you can start APC-cy7 or APC with bivariate histograms (height vs area). Even you tried with all series of reasonable numbers, you might miss another things. For example, Even though you have a right number for laser delay, your sample population is located in out of range with 'high enough' PMT voltage (Sometimes people only seek for the answer they want.). Thus, you can find this population with some tricks. First, you must have good FSC vs SSC signals. Second, you can gate previous main population into APC-cy7 or APC histograms. (sure, you can not see anything.) Third, you can make box (or gating box) inside APC-cy7 or APC histograms and try to find exact number of cells, which are gated from FSC vs SSC histogram by moving it around from corner to corner. You can see the changes of number for the population in statistics box. Fourth, if you find it, try to put those cells in the diagonal line of APC-cy7 or APC bivariate histograms (height vs area) with fine tuning. Hope it helps, Chang % Do not forget checking other basic stuffs. (wrong beads etc...) -----Original Message----- From: Achal Pashine [mailto:achalpashine@gmail.com] Sent: Tuesday, April 05, 2005 2:55 PM To: Cytometry Mailing List Subject: FACSAria Question: no signal with Red laser (Trigon) Dear Flowers, This is Aria related question and any tips would be greatly helpful to us (till BD guys get back to us). Until recently, when we ran QC with rainbow particles, we could see signals in Blue laser related PMTs (octagon) as well as Red lazer related PMTs (trigon). But today when I ran QC, I do not detect any major signal on Red related PMTs with the same batch of rainbow beads. We checked (and incrementally changed) laser delay from -30 to -50 and did not get any signal. Also, changed area scaling (from .5 to 2.0 incremental changes) without any effect. (also changed sheath pressure:low high med as well as checked instrument configuration to see whether red related PMTs are included or not, they are). I phyisically checked red laser is working by using shutter override and can see both the colors (blue and red). Curiously, we see some very sharp noise like signal at ~60 in APC-cy7 as well as APC area histogram that does not move at all with change in voltage of that PMT. Do you have any tips as to how to regain signal acquisition for our problem or what can be wrong. Thanks in advance for your insights. -- Achal Pashine Chiron CorporationReceived on Thu Apr 7 18:42:59 2005
This archive was generated by hypermail 2.1.8 : Sat Jan 14 2006 - 22:03:42 EST