RE: FACSAria Question: no signal with Red laser (Trigon)

From: Loh Siew Chin <Loh_Siew_Chin@bsf.a-star.edu.sg>
Date: Wed Apr 06 2005 - 21:08:56 EST
Hi Achal,
 
     Personally, I also have the problem to adjust laser delays.
Sigh.... Not very good in that. Your problem could be due to your red
laser alignment, I guess. If that's really the case, then you gotta wait
for ur BD guys to help. Another posibility of not seeing any signals,
could be your cell flow is 'dirty'. But this is quite unlikely if
signals with the other 2 lasers can been detected. 
 
Cordially,
 
Siew Chin
 
 

  _____  

From: Achal Pashine [mailto:achalpashine@gmail.com] 
Sent: Wednesday, April 06, 2005 3:55 AM
To: cyto-inbox
Subject: FACSAria Question: no signal with Red laser (Trigon)


Dear Flowers,
This is Aria related question and any tips would be greatly helpful to
us (till BD guys get back to us). 
Until recently, when we ran QC with rainbow particles, we could see
signals in Blue laser related PMTs (octagon) as well as Red lazer
related PMTs (trigon). But today when I ran QC, I do not detect any
major signal on Red related PMTs with the same batch of rainbow beads.
We checked (and incrementally changed) laser delay from -30 to -50 and
did not get any signal. Also, changed area scaling (from .5 to 2.0
incremental changes) without any effect. (also changed sheath
pressure:low high med as well as checked instrument configuration to see
whether red related PMTs are included or not, they are).
I phyisically checked red laser is working by using shutter override and
can see both the colors (blue and red).
Curiously, we see some very sharp noise like signal at ~60 in APC-cy7 as
well as APC area histogram that does not move at all with change in
voltage of that PMT.
Do you have any tips as to how to regain signal acquisition for our
problem or what can be wrong.
Thanks in advance for your insights.

-- 
Achal Pashine
Chiron Corporation
Received on Thu Apr 7 18:38:00 2005