Dear Ioannis, Concerning your questions: 1.I never performed ethanol fixation after APC staining, but ethanol might affect APC, just like it affects R-PE, published 12 years ago by Ken Toba in Cytometry 1992;13(1):60-7. "Improved staining method for the simultaneous flow cytofluorometric analysis of DNA content, S-phase fraction, and surface phenotype using single laser instrumentation." We observed exactly the same. 2. It is no problem using log amps for FITC and APC. 3. It might be possible that you are dealing with doublets. For a more optimal discrimination between singlets and doublets using PI and APC as fluorescent dyes for labelling DNA and a protein, respectively, you need to change the hardware of the Calibur, since APC is measured on the FL4 at the costs of the pulse-width (FL1, FL2 or FL3). The solution is quite simple. You only need a new connecting wire (can be obtained from BD) and make some disconnections and new connections in the Calibur. The protocol was published in the following paper: "Four-color multiparameter DNA flow cytometric method to study phenotypic intratumor heterogeneity in cervical cancer. Cytometry. 2000 Feb 1;39(2):96-107." Good luck, Willem Corver -----Original Message----- From: Kotsianidis, Ioannis [mailto:i.kotsianidis@imperial.ac.uk] Sent: maandag 20 december 2004 16:03 To: cyto-inbox Subject: SIMULTANEOUS SURFACE AND PI STAINING QUESTION Dear flowers, I have a couple of questions regarding simultaneous surface and cell cycle analysis with PI. I stained a lymphoblastoid cell line with FITC and APC conjugates, subsequently fixed it with 70% ethanol and then incubated it with RNAse and PI. I acquired the cells in a FacsCalibur (settings set for cell cycle analysis, PI is read on FL2-A) and I saw a strange picture of double positive events at the FL1 vs FL4 plot. I repeated the acquisition with my usual four color settings and I've noticed that there was no positivity for APC, while my unfixed sample was all double positive for both FITC and APC. 1) Does the fixation destroyed the APC epitope (if yes, then why not also the FITC)? 2) Can I use FL4 and FL1 in Log scale for acquisition and analysis, while I perform cell cycle analysis with PI? Thanks in advance, Dr Ioannis Kotsianidis Clinical Research Fellow Department of Haematology Faculty of Medicine - ICSTM Hammersmith Hospital Du Cane Road LONDON W12 0NN U.K.Received on Tue Dec 21 12:58:00 2004
This archive was generated by hypermail 2.1.8 : Wed Dec 22 2004 - 03:12:05 EST