Re: Vertical Motion timeouts on Facscalibur

From: Raymond Lannigan <rlannigan@cytekdev.com>
Date: Wed Dec 08 2004 - 20:09:23 EST
Kate,
	You have found the achilles heel of this instrument. Sounds like your 
worm gear, and other lifter components needs a good cleaning along with 
some light lubrication. It's been my experience that Teflon type 
lubricants work best. After changing the lifter motor, there are some 
adjustments that can be tricky for someone trying to do this for the 
first time. There are also some settings in the Loader manager software 
which may be adjusted to the lifter parameters. While I am eager to 
visit your part of the world, it might be easier for you to send us the 
loader drawer for repair. If you would like more info. please feel free 
to contact me.
Ray
Raymond Lannigan
Cytek Development
19331 Briar Patch Dr.
Gordonsville, VA 22942
540-832-2793
540-832-2794 (fax)
434-531-3607 (cell)

On Dec 7, 2004, at 7:57 PM, Kate Marson wrote:

> Hi all
>
> We have been experiencing intermittent problems with the loader on our 
> BD Facscalibur.  We get numerous 'Vertical Motion timeouts' whilst 
> trying to run samples on Worklist Manager.  This locks up the system 
> and requires us to run manually, which is time consuming.  Our patient 
> service engineers have replaced the motor inside the loader and then a 
> week later when that hadn't fixed things, replaced the entire loader 
> mechanism.  10 days later we have the same problems as before, but 
> with slow vertical motion instead of the rigid jamming that we had 
> before the loader was replaced.  Has anyone else experienced similar 
> problems with their loader, and could someone please let us know what 
> you have had done to fix this…
>
> Many thanks
>
> Kate Marson
>  Technical Specialist
>  Diagnostic Immunophenotyping
> Haematology
>  Labplus, ADHB,
>  Auckland,
>  NZ. 
>