A combination re the above configuration that would detect APC-Cy7, 
APC-Cy5.5 and APC is as outlined below

APC-Cy7       (emission 780) Trigon position A	Dichroic  755LP  Band pass 
  780/60
APC-Cy5.5    (emission 695) Trigon position B  Dichroic  675LP	Band pass  
690/15
APC		   (emission 660)  Trigon position C   Dichroic Blank Band 
pass  660/20


	The reduction of the APC-Cy7  signal needs to be checked out , by 
what criteria are you saying the signal is reduced , also to what 
magnitude ?
A nice way to evaluate the signal detection for each of the PMT's is to 
use some Comp Beads with the relevant fluorochrome on each one then run 
them and verify the detection of signal in each PMT, it is a good simple 
way to start .


Ian 