RE: Broken connector on FACScan [TS-EQ84C14]

From: David Coder (d_coder@MSN.com)
Date: Tue Dec 24 2002 - 10:31:32 EST


Tony,

As I recall the connector leads to a filter for venting. If (and only
if) you are running samples that pose no biohazard, you can easily run
the instrument for the short term by simply removing the unbroken end of
the quick-disconnect coupler and allowing the waste tank to vent to the
air. (Again, for completely safe and non-biohazardous samples, you could
simply run the waste line from the instrument straight into the waste
water drain in lab. The waste tank will never get full.)

The couplers are made by Colder Products Co. (www.colder.com). Get
several as the bulkhead couplers seem to break more frequently (it's
probably the angle at which they're remove/inserted--unpredicted design
'feature').  For the waste connectors, you could avoid any future
breakage by getting a chrome-plated brass coupler; more expensive but it
won't break.

Dave
------------------
David M. Coder, Ph.D.
Consultant in Cytometry
Seattle, Washington
tel./message: 206-499-3446
email: d_coder@msn.com


-----Original Message-----
From:
Sent: Saturday, December 21, 2002 5:01 AM
To: cyto-inbox
Subject: Broken connector on FACScan


The white waste container connector on our FACScan has broken and I'm in
need of some assistance. This is the connector that's attached to the
instrument, not the tank (see attached photos). The connector looks as
though it's supposed to be connected to a line that runs into the
instrument, but I cannot see one. Anyone know about this? Also, what
does one of these things cost to replace?

Thanks,

Tony
--
Tony Schountz, Ph.D.
Dept. of Biology
Mesa State College
Grand Junction, CO  81506
http://www.mesastate.edu/~tschount/