Re: MacOS 9 and CellQuest

From: Pál Jáksó (Pal.Jakso@aok.pte.hu)
Date: Fri Dec 20 2002 - 03:35:12 EST


Hi Ken,

I've tried OS 9.1 and yesterday upgraded to 9.2.1 but it did not help.
I'm little bit confused because of the many different advices from this
mailing list.
Somebody says that they can work with Cellquest perfectly using the new
versions of OS9 ignoring the BDPAC opening problem. But without this I
can not adjust the detector settings.
Many of them say that I should use the BDInits installer 3.4. Yes I use
exactly this version and still does not work.
A third opinion suggest to use BDInits installer 3.5 - I'm waiting for
that from my local BD representative.
The dongle what I use is a regular ADB dongle resting still in the same
port where it work under OS 8.6.
I don't intend to use USB dongle, the FireWire/USB card is needed only
for some other USB devices like printer, USB FlowJo dongle and FireWire
CD writer.
And what is that key combo to BDPAC ?
Otherwise thinking logically my problem with opening BDPAC may
originate from some type of conflict between the OS and this version of
BDPAC. If I had problem with the dongle I guess I could not open
CellQuest but I can(!) just the aquiring functions are affected.

Thank you and everybody who tried to help me, Merry Christmas and Happy
New Year!

Pal

On Thursday, Dec 19, 2002, at 17:13 Europe/Budapest, McDonald, Ken
wrote:

> Hi Pal,
>	Are you using the newest MacOS 9? I've used CellQuest with OS 9.04
> and it worked fine. I would use the older version of MacOS 9. Also
> make sure
> the dongle that comes with the program is right for the USB/Firewire
> card
> you have.You still have to use the key combo to access BDPAC...Ken
>
> Ken McDonald (Biologist)
> Merck Frosst Canada
> Tel: 514-428-2802
> email: ken_mcdonald@merck.com
> www.merckfrosstlab.ca
>
>
> -----Original Message-----
> From: Pál Jáksó [mailto:Pal.Jakso@aok.pte.hu]
> Sent: Wednesday, December 18, 2002 9:30 AM
> To: Cytometry Mailing List
> Subject: MacOS 9 and CellQuest
>
>
>
> Hello Flowers,
>
> Until recently  we used MacOS 8.6 with CellQuest without any problem
> but we've bought a FireWire/USB combo card for external devices and it
> requires MacOS 9. My problem is that under OS9 I can't configure the
> BDPAC control panel because of an error  - The application "BDPAC"
> could not be opened because "BDPAC" could not be found (-2,806). The
> installation read me says that I should open BDPAC while holding down
> the command and control keys but it does not help.
> Anybody have an idea ?
>
> Thanks in advance,
>
> Pal Jakso
>
> University of Pecs
> Faculty of Medicine
> Dept. of Pathology
> 7643, Hungary
> 12. Szigeti str.
>
>
> -----------------------------------------------------------------------
> -------
> Notice: This e-mail message, together with any attachments, contains
> information of Merck & Co., Inc. (Whitehouse Station, New Jersey, USA)
> that may be confidential, proprietary copyrighted and/or legally
> privileged, and is intended solely for the use of the individual or
> entity named on this message. If you are not the intended recipient,
> and have received this message in error, please immediately return
> this by e-mail and then delete it.
>
> =======================================================================
> =======
>


This archive was generated by hypermail 2.1.6 : Thu Jan 01 2004 - 17:42:13 EST