RE: Gating apoptotic sub-G1 cells

From: DARZYNKIEWICZ ZBIGNIEW (DARZYNK@nymc.edu)
Date: Wed Dec 04 2002 - 18:38:06 EST


The gating strategy for sub-G1 cells is described in detail in  Meth Cell
Biol.,Vol  63: pp 533-4, 2001 and also in the review in Cytometry, 27, 1-20,
1997. Unfortunately, many authors that use this methodology, use it
erronously. In short, these are the critical points.

1. One should not lyse unfixed cells using detergents or hypotonic
solutions. Because nucleus of apoptotic cell is fragmented, during cell
lysis many nuclear fragments are released from a single cell.  Most authors
count these fragments as individual apoptotic sub G1 cells. This
overestimates apoptotic index (AI) many-fold.In addition, lysis of mitotic
cells releases individual chromosomes or chromosome aggregates - each of
them is erronously counted as an apoptotic cell. If the cells were treated
with taxol or other mitotic poisons, this can result in incredibly high
error !!!! Micronuclei also are released and also are erronously counted as
apoptotic cells.

2. One should not use logarithmic scale for ploting intensity of
DNA-associated fluoprescence. Note that when using log scale many authors
identify and count the objects with DNA content less than 1 % ,and some
authors with less than 0.1% of that of the whole G1 cells, as apoptotic "sub
G1" cells. The objects with so little DNA content cannot be the whole
apoptotic cells. Most  likely they are small chromatin (nuclear) fragments
or other objects discussed in point #1.

3. I suggest that the cells have to be fixed in ethanol and then extensively
rinsed, pehaps even with high molarity buffer  to extract fragmented DNA,
prior to staining. Then the sub-G1 cell fraction should be identified
(counted) by setting the low level gate at 5 or 10 % value of the G1 peak.
Thus, if the peak of G1 cells is at channel 200, the objects with
fluorescence lesser than 10 or 20 channels should be excluded. The 5 or 10 %
level is, of course, arbitratry and may lead to understimation of apoptotic
cells. The degree of underestimate, however, is consistent and reproducible
from sample to sample.

Zbigniew Darzynkiewicz, M.D., Ph.D.

New York Medical College
19 Bradhurst Ave.
Hawthorne, NY 10532
tel: 914-347-2801
fax: 914-347-2804
http://www.geocities.com/z_darzynkiewicz



> -----Original Message-----
> From: Inna.Cohen@med.va.gov [SMTP:Inna.Cohen@med.va.gov]
> Sent: Monday, December 02, 2002 2:50 PM
> To:	Cytometry Mailing List
> Subject:	Gating
>
>
> Hello All!
> We at the VAMC are having problems deciding where to gate to see a subG1
> peak when we look for apoptosis in the cell cycle.  Any helpful
> suggestions?
>
> Thanx,
> Inna Cohen
> VA Medical Center


This archive was generated by hypermail 2.1.6 : Thu Jan 01 2004 - 17:42:12 EST