Re: [7AAD viability for CD34]

From: Ian_DIMMICK@europe.bd.com
Date: Wed Apr 10 2002 - 02:47:45 EST


I have used 7-AAD for some time now for this application and I find the
following rules very helpful

1    Remember that 7-AAD is also capable of being absorbed into live as
well as dead cells , the uptake into dead is rapid , but into live much
slower ,       however to get maximal differentiation from live to dead
analyse as soon as you can after the initial antibody incubation period is
complete
2    Be careful which Lysis system you use, some will permeabilise slightly
and consequently can not be used with this dye, try to use an ammonium
chloride  based lysis
3    7-AAD is quite bright to the point it may start to quench some
detectors , do not apply ver y high volts/gains , it is not necessary and
may cause you to have    to run with high compensation needlessly
4    I use 5ul of 7-AAD in with the 45/34 cocktail , this works well (using
7-AAD from BD-Pharmingen)


Hope this is helpful

Ian